Dear guru,

I have 3 questions;

Background: an inspection lot created with two inspection specification. The first inspection characteristic has a control indicator of summarize recording while the second inspection characteristic has a control indicator of single recording.

1st Question:

The first inspection characteristic has a mean value (i.e. value in the result field) which fall outside the lower and upper limit, but system valuated it as ACCEPT (i.e. accept box checked, and reject box unchecked, and the valuation field has a green check sign). Why? Shouldn't it be rejected as the result (i.e. mean value) fall outside the range?

I notice the problem not only incurs to inspection characteristic with summarize recording, but single recording as well.

2nd Question:

The second inspection characteristic (which is single recording) has ONE actual sample size that is tested fall above the Upper limit.

Out of the total 125 sample sizes value entered, one sample size value is falling outside the upper limit (the defect class is u201C01 Critical defectu201D and the u201CNo. Aboveu201D field is 1), and the system consider the this inspection characteristic as u201CAcceptu201D (i.e. the accept checkbox is checked and reject checkbox is uncheck when I hits the Close (i.e. lock sign) button; Further the valuation field is checked. Shouldnu2019t the system evaluate it as reject (i.e. with the rejected box checked and accept box uncheck if one of the inspection sample size has a value which is either fall outside the upper or the lower limit?

3rd Question:

Do I always have to press the Close (i.e. lock sign) button in order to trigger the acceptance and rejection (i.e. the two checkboxes) of the inspection characteristic?

Thanks.

YY

View Entire Topic
Former Member
0 Kudos

Hi Gajesh and Sujit,

1st question u2013 May I know under what combination of <Sampling Type and Valuation Mode> the system will use the Non-Conforming Units (i.e. Number of Rejected Unit and Accept Unit in the sampling plan table) for valuation and under what combination is the valuation is on Mean value?

The following are the valuation mode on standard SAP:

100 u2013 attribute inspection nonconf. Unit

110 u2013 attribute inspection nonconf. Unit / Manual

200 u2013 attribute inspection number of defect.

210 u2013 attribute inspection number of defect /Manual

300 u2013 variable inspection s-method (one limit)

310 u2013 variable inspection s-method (two limits)

400 u2013 valuation according to character attribute code

500 u2013 manual valuation

510 u2013 without valuation parameters

700 u2013 mean value within tolerance range

800 u2013 SPC inspection.

3rd question u2013 The result configuration indicator used in the plant level setting is empty (i.e. no background function). That explains why I need to manually clicks the Lock icon button whenever I perform result recording on this plant.

Thanks

YY.

Former Member
0 Kudos

Dear YY,

You nee to select Sampling Type as per your reqmnt and accordingly if you select "700 Mean Value within tolerance range" system automatically valuates the characteristics based on the Mean Value.

Best Regards,