SAP EAM lack a metrology platform for the calibration of lab and process instrumentation.

I have seen that many manufacturing companies are not using SAP EAM for their labs and process calibration programs. Where measurement values are compared with those of a calibration standard, and the "as found" are recorded, and the after calibration is performed then the "as left" are also recorded all of this following a given calibration frequency, and as found and as left reading should be within an applicable tolerance. Now, as far as any of you know, does SAP have at least basic metrology functions that can be used to track calibration performance?

Add comment
10|10000 characters needed characters exceeded

  • Get RSS Feed

7 Answers

  • Nov 28, 2016 at 02:33 PM

    Edgar,

    What values do you need to record and compare?

    PeteA

    Add comment
    10|10000 characters needed characters exceeded

  • avatar image
    Former Member
    Nov 28, 2016 at 04:31 PM

    For example:

    I might have a pressure transmitter with a range of 0-200 PSI, and need to check the deviation at 3 point in the range. Let say the nominal points are 50, 100 and 150 with a tolerance of +/- 5% of the reading.

    In this case, I would have the 'As Found" readings and after calibration if needed, then I will have the "As Left" readings.

    Reading as found

    Add comment
    10|10000 characters needed characters exceeded

  • avatar image
    Former Member
    Nov 28, 2016 at 04:31 PM

    For example:

    I might have a pressure transmitter with a range of 0-200 PSI, and need to check the deviation at 3 point in the range. Let say the nominal points are 50, 100 and 150 with a tolerance of +/- 5% of the reading.

    In this case, I would have the 'As Found" readings and after calibration if needed, then I will have the "As Left" readings.

    Nom. Pt. ....Reading as found-----As found deviation......Reading as left------As left deviation

    50.....................46...............................-8%.................................50...............................0

    100...................106.............................+6%................................101.............................1%

    Hope you got the idea.

    Add comment
    10|10000 characters needed characters exceeded

  • Nov 28, 2016 at 05:01 PM

    Edgar,

    This would require PM/QM integration using inspection lots and inspection results.

    http://help.sap.com/saphelp_pserv464/helpdata/en/48/e82c7dd435d1118b3f0060b03ca329/frameset.htm

    PeteA

    Add comment
    10|10000 characters needed characters exceeded

  • avatar image
    Former Member
    Dec 22, 2016 at 01:15 PM

    Edgar,

    I understand what you mean and it's not very straightforward in SAP. But with some workaround this can be achieved. You need to create MIC's as below

    1. Nom. Pt. Reading at 50 Specification( 0-200 PSI) Result=50

    2. Reading as found at 50 Specification( 0-200 PSI) Result=46

    3. As found deviation at 50 Specification( +/- 5%) Result= (50-46/50)*100

    MIC number 3 will be calculative MIC. In the same way create MIC's for reading at 100 & 150.

    Thanks

    Add comment
    10|10000 characters needed characters exceeded

    • Former Member

      Gupta:

      Could you shed more light on how to create a MIC. I haven't heard of MIC, is this a Tcode. Or do I need a custom made template?.

      Thank you.

  • avatar image
    Former Member
    Jan 17, 2017 at 05:17 AM

    Edgar,

    MIC creation is part of QM module, please check with your QM consultant on the same.

    Thnaks

    Add comment
    10|10000 characters needed characters exceeded

  • avatar image
    Former Member
    Jan 23, 2017 at 06:40 PM

    Gupta:

    Thank you for the information.